  • @robbrown2007
    @robbrown2007 Год назад +1

    I would respectfully disagree that calibration includes adjusting an instrument. Determining and applying correction factors yes, but adjustment no.

    • @olewurtz7625
      @olewurtz7625 Год назад +2

      What is the difference between correction factor and adjustment ?
      In my understanding, adjustment is a physical manipulation of say a resistor or capacitor to obtain a desired readout or output from the DUT. A correction factor, a computer applied correction value stored in the DUT’s memory.
      So in principle the same process used on typically older and newer devices.
      And, a calibration is a verification of specifications, not necessarily including a manipulation of correction input.

    • @Vintaginside
      @Vintaginside Год назад +2

      You can even adjust an instrument every single time it goes for calibration. The key factor is giving the cal certificate with measured values before and after adjustment.